Hello All,
I am trying to find the 3D pressure field generated by an ultrasound transducer by using the laser vibrometer data (velocity) measured on the transducer surface as the source. I have set source.u_mode = 'dirichlet'
, however, after going over this post and this example which is about recording and re-transmitting a signal using a Dirichlet BC, I have noticed that my results may be incorrect.
The transducer had a lens on it and the measurements where done in a water tank. So, as also discussed in the post above Dirichlet BCs are useful when the medium is homogeneous. I actually tested this in the example above and max percentage error increases from %1-2 to around %20 when I split the medium into 2 different media. As far as I know I cannot utilize the Equivalent-Source Acoustic Holography method since it requires measured pressure data instead of velocity.
What would be the best approach for my problem?
Thanks,
Khan