Hi george020106,
As Brad said in his reply, the shape of the pulse envelope will depend on the transducer bandwidth. Often the ring up and ring down are a few cycles each for a fairly broadband piezoceramic transducer. I don't have a specific reference but it is a fairly commonly used approximation. As you have experimental data then you can verify if this is true in your case.
For toneburts of longer than a few cycles a guassian envelope is not very suitable. For many transducers there will be a ring up time which could be up to 8 to 10 cycles in some cases and a ring down of a few cycles at the end (depending on how resonant the transducer is) with a stable amplitude in between. There is some discussion of this in the literature and International standards (IEC 62556 on HITU field characterisation) related to HIFU transducers. You could look up the following references, I'm sure there are others also:
Khokhlova et al. 2013 J. Acoust. Soc. Am. 134 (2); 1586.
Jenderka and Wilkens. 2012 Metrologia, 49 (2012) S275–S278.
There will be an updated version of the toneburst function in the next release to allow you to choose the number of ring up and ring down cycles or change the shape of the envelope.
Elly